I recently was looking at specifications for solar powered vaccine refrigerators in the PIS and PQS specifications. I was trying to do a comparison of the various models and found it difficult.
For each model the data given is for different test conditions. For example, energy use data for True Energy Model PS65i is given for stable run conditions at 43 deg C, 32 deg C, and 27 deg C and also energy use for the cool down test at 43 deg C. No energy use figures are given with ice making. For the Dometic SARL TCW 2000DC 43 deg C energy use is given for stable run conditions with the refrigerator compartment on and freezer turned off and energy consumption for ice making at 43 deg C.
When sizing a solar system the data most commonly needed is energy consumption at 32 deg C with and without ice making.
Another point which could be presented more clearly is which of the compressors is running on two compressor models.
These issues could easily be cleared up; since all the necessary data was generated when the PQS tests were carried out.