Post00314 VVM - DRY ICE TROUBLE? NO! 25 January 2001
CONTENTS
1. VVM on OPV: DRY ICE TROUBLE? NO!
This discussion revolves around misinformation concerning VVMs.
Fred Simiyu, WHO Nigeria, reported "Last week I heard it mentioned by a
UNICEF colleague that a recent study indicates dry ice, when used to pack
vaccines with VVMs, causes change in VVM colour simulating exposure to
heat."
Fred sent an email to Modibo Dicko, WHO/AFRO, enquiring if VVMs had a
problem with dry ice exposure, and wondering if that would account for VVM
changes observed in a recent oral polio vaccine shipment to one district.
Hans Everts, WHO/EPI, initially responded, pointing out that the heat
exposure may have occurred prior to being packed for shipment.
Alissa Klopper, Lifelines Inc, the VVM company, reports on the testing of 4
production lots from different manufacturers, showing no change on exposure
to dry ice - for as many as 8 days.
* The answer to Fred is that the problem was that heat exposure caused the
VVMs to change.
Thanks to Julie Milstien, V&B/ATT and Sara Stappers, V&B/ATT, for posting
the discussion. Thanks also to the Fred, Hans, and Alissa for generating a
rapid clarification of puzzling field issue.
Opinion, comments and additions please: [[email protected]][email protected][/email]
or use your reply button
___________________________________________________________________________
From: [[email protected]][email protected][/email]
To: [[email protected]][email protected][/email]
Subject: FW: Dry Ice Trouble???
Date: Wed, 17 Jan 2001 10:34:51 +0100
Dear Allan,
Please find below details of correspondence regarding a problem with VVMs in
Nigeria. Julie Milstien asked me to contact Lifelines Technology Inc. for
their response, which I have enclosed here.
Dr Milstien also asked me to request that this information is posted to
Technet, I would be very grateful if you would do this for us.
Please don't hesitate to let me know if you should need any further
information.
Best regards,
Sara Stappers
Sara Stappers
Secretary
ATT, V&B
World Health Organization
Tel. (41) 22 791 2051
Fax (41) 22 791 4384
___________________________________________________________________________
Subject: Dry Ice Trouble???
Author: "Fred Simiyu"
Date: 11/29/00 10:26 PM
Dear Modibo,
Last week I heard it mentioned by a UNICEF colleague that a recent study
indicates dry ice, when used to pack vaccines with VVMs, causes change in
VVM colour simulating exposure to heat. Probably a chemical result of the
sublimation process? Is anyone out there aware of this study? Is it
documented?
During the just concluded R2 of NIDs, OPV was transported to one of our
States, Anambra. NIDs OPV was packed with dry ice while routine OPV was
packed with conventional ice packs. All the OPVC had VVMs at stage 1 on
departure. The truck had some trouble on the way and arrived over 15 hrs
later. On arrival, all OPV for NIDs were at VVM 2 while that for routine
was at VVM 1. Another factor that could have caused this was that the
routine OPV was packed with more ice packs than that for NIDs due to a
number of reasons that we have since addressed here.
Could it also have been 'dry ice trouble'?
Regards,
Fred.
---
From: [[email protected]][email protected][/email]
Sent: 02 December 2000 11:45
To: Fred Simiyu
Subject: Re:Dry Ice Trouble???
Hi Fred!
I don't have an off-the-shelf answer to this question and if Hans does not
either, then it would be good to set up a small experiment scientifically
organized. What do you think?
Modibo
---
Subject: RE: Dry Ice Trouble???
Author: evertsj at whohq6
Date: 04/12/2000 16:09
Interesting problem. I do not know of studies regarding the effect of dry
ice on VVM. Even with a complete breakdown of the cold chain, if the vaccine
was packed properly from the start and if the VVM were at 1, the VVM should
only slightly have started to change colour.
Even though all VVM looked like being at step 1, they may have had different
heat exposure before that. In the beginning the difference in colour may not
be very obvious. Do you know when the batches arrived and where they were
stored.
Hans Everts
Technical Officer
EPI
WHO Geneva
Tel: 00 41 22 791 3683
---
From: [[email protected]][email protected][/email]
Sent: Monday, 4 December 2000
Subject: RE: Dry Ice Trouble???
Sara, could you please forward to Ted Prusik, Lifelines, for response
___________________________________________________________________________
* TEXT OF LIFELINES TECHNOLOGY INC. RESPONSE
___________________________________________________________________________
LifeLines Technology, Inc.
116 American Rd.
Morris Plains, NJ 07950
USA
Tel: (973) 984-6000
Fax: (973) 984-1520
January 2, 2001
TO: Sara Stappers
CC: J. Milstein, T. Prusik
Dear Ms. Stappers,
Tests have been conducted on four lots (different manufacturers) of OPV VVMs
to verify that dry ice has no effect on the reactivity of VVMs. The table
below shows that the OD does not change during six days of frozen storage
with dry ice, blue ice packs, or in a conventional freezer.
OD(reference) - OD(indicator)
Lot Number
Days of Storage Blue ice packs Dry ice -30C freezer
JG343/1
0 0.37 0.37 0.37
2.1 0.36 0.36 0.37
5.8 0.37 0.37 0.37
JG286/1
0 0.32 0.32 0.32
2.1 0.33 0.32 0.33
5.8 0.33 0.33 0.33
JG358/1
0 0.35 0.35 0.35
2.1 0.35 0.35 0.35
5.8 0.36 0.36 0.36
JG349/1
0 0.35 0.35 0.35
2.1 0.35 0.35 0.35
5.8 0.35 0.36 0.36
Our conclusion is that the OD issue identified in Anambra is not due to dry
ice.
Sincerely,
Alissa Klopper
QA Manager
____________________________________*______________________________________
There are no replies made for this post yet.